Polydoros IT Error 045

Tags: Radiology - R/F, SIEMENS

sohail aftab

Polydoros IT Error 045

May 04, 2012 09:39

Dears all help me. i have a problem in D500 PCB, siemens model(Polydoros IT) generator according to Error 045 i want help regarding this problem.

Err045: High Voltage Cable Configuration.

Description:
• Configuration of the high voltage cable length must be performed.
Possible Causes of Error:
• System configuration not correct.
• D500 defective.
Corrective Measures:
• Correctly configure the high voltage cable lengths
• If the error occurs again following reconfiguration, replace the D500.

i also check all thinks according to Error Description all is OK with out D500 Please help me how to conform D500 is not proper working.
I'm waiting,
supremeditech@yahoo.com