Dears all help me. i have a problem in D500 PCB, siemens model(Polydoros IT) generator according to Error 045 i want help regarding this problem.
Err045: High Voltage Cable Configuration.
Description:
• Configuration of the high voltage cable length must be performed.
Possible Causes of Error:
• System configuration not correct.
• D500 defective.
Corrective Measures:
• Correctly configure the high voltage cable lengths
• If the error occurs again following reconfiguration, replace the D500.
i also check all thinks according to Error Description all is OK with out D500 Please help me how to conform D500 is not proper working.
I'm waiting,
supremeditech@yahoo.com Type your question or comment here...
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
sohail aftab
Polydoros IT Error 45
May 05, 2012 04:22
Dears all help me. i have a problem in D500 PCB, siemens model(Polydoros IT) generator according to Error 045 i want help regarding this problem.
Err045: High Voltage Cable Configuration.
Description:
• Configuration of the high voltage cable length must be performed.
Possible Causes of Error:
• System configuration not correct.
• D500 defective.
Corrective Measures:
• Correctly configure the high voltage cable lengths
• If the error occurs again following reconfiguration, replace the D500.
i also check all thinks according to Error Description all is OK with out D500 Please help me how to conform D500 is not proper working.
I'm waiting,
supremeditech@yahoo.com Type your question or comment here...
to rate and post a comment