Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Polydoros sx 65 doesn´t start up

Tags: Radiology - R/F, SIEMENS

(53)
(9)

EFREN FRANCO

Polydoros sx 65 doesn´t start up

June 25, 2013 08:11

I have a problem with a Polydoros sx 65. It doesn´t complete the start up process. The display in D100 remains in "1" (waiting for network connection) and the leds indicators SPC "tx rx" in the XCU doesn´t blink. Does anyone have an opinion about this problem?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Polydoros sx 65 doesn´t start up

June 26, 2013 10:10

sounds like a XCU issue possibly bad hard drive. do you have a a XCU boot disk you can see if the system will bootup with floppy in the drive?

Log inor Register

to rate and post a comment

(192)
(45)
(1)

Dante Nuevo

re: Polydoros sx 65 doesn´t start up

June 26, 2013 10:10

Contact me. I can help you.
Dante Nuevo

dantenuevo@hotmail.com

Log inor Register

to rate and post a comment

(56)
(27)

ADRIAN GUTIERREZ

re: Polydoros sx 65 doesn´t start up

June 26, 2013 10:17

look for some info on smstechforum.com they help me a lot.

Log inor Register

to rate and post a comment

(53)
(9)

EFREN FRANCO

re: re: Polydoros sx 65 doesn´t start up

June 27, 2013 12:40

thank you for the information.

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: Polydoros sx 65 doesn´t start up

June 26, 2013 10:28

Terminate communication ports with 120 Ohms resistors on D320 and see if the generator becomes ready. This way you understand if error comes from external device.
Also, you can connect monitor and keyboard to the XCU and see if the pc is booting normally. There is a battery also in the XCU which must be replaced regularly.

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: Polydoros sx 65 doesn´t start up

June 26, 2013 10:39

The ready state of D320 is 0.0. and the decimals must flash if i remember well.Other states have error explanation as well.
If you replace CMOS battery you must configure the XCU CMOS >boot without keyboard

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: Polydoros sx 65 doesn´t start up

June 26, 2013 10:39

The SPC tx/rx is indicative when you connect service PC with the siemens software.

Log inor Register

to rate and post a comment

(53)
(9)

EFREN FRANCO

re: re: Polydoros sx 65 doesn´t start up

June 27, 2013 12:40

thank you for your reply. I'll try to find out something with these comments.