Meaning:
Dose monitoring has responded, not enough dose detected after 100 ms
Description:
In an iontomated exposure, the dose counter on D100 is read after 100 ms exposure time. If one half of the necessary does is not reached by the max. exposure time, error 550 is signalled.
Possible causes and action:
Faulty operation (customer):
Tube not directed onto selected IONTOMAT chamber
Collimator closed
Wrong choice of the exposure kV
Errors:
Wrong IONTOMAT workstation programmed, check system configuration
No signal from IONTOMAT chamber at measuring point D100.X63 DL_IN
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Charbel Tadros
Polydoros sx80 error 550
January 15, 2009 01:33
the generator freeze and give error 550 after the exposure. Need help ..
Urgent
to rate and post a comment
Henrique Gomes
Re: Polydoros sx80 error 550
January 16, 2009 09:39
Meaning:
Dose monitoring has responded, not enough dose detected after 100 ms
Description:
In an iontomated exposure, the dose counter on D100 is read after 100 ms exposure time. If one half of the necessary does is not reached by the max. exposure time, error 550 is signalled.
Possible causes and action:
Faulty operation (customer):
Tube not directed onto selected IONTOMAT chamber
Collimator closed
Wrong choice of the exposure kV
Errors:
Wrong IONTOMAT workstation programmed, check system configuration
No signal from IONTOMAT chamber at measuring point D100.X63 DL_IN
Check detector and cabling
D100 defective, replace
Sincerelly,
gomes : neutrinno@zipmail.com.br
to rate and post a comment