Polydoros sx80 error 550

Charbel Tadros

Polydoros sx80 error 550

January 15, 2009 01:33

the generator freeze and give error 550 after the exposure. Need help ..
Urgent

Log inor Register

to rate and post a comment

(120)
(15)
(15)

Henrique Gomes

Re: Polydoros sx80 error 550

January 16, 2009 09:39

Meaning:
Dose monitoring has responded, not enough dose detected after 100 ms


Description:
In an iontomated exposure, the dose counter on D100 is read after 100 ms exposure time. If one half of the necessary does is not reached by the max. exposure time, error 550 is signalled.


Possible causes and action:
Faulty operation (customer):
Tube not directed onto selected IONTOMAT chamber

Collimator closed

Wrong choice of the exposure kV

Errors:
Wrong IONTOMAT workstation programmed, check system configuration

No signal from IONTOMAT chamber at measuring point D100.X63 DL_IN

Check detector and cabling

D100 defective, replace


Sincerelly,

gomes : neutrinno@zipmail.com.br