Your Email:
preecha chaipet
polymobile 3 error E14
March 28, 2011 10:04
hello anybody i have problem with sement polymobile 3 rotor okay but can't exposure with code error E14 what does it error mean ? regards
to rate and post a comment
Alfonso Nino
Re: polymobile 3 error E14
March 29, 2011 07:42
kV-ACT
March 31, 2011 08:20
Hi friend thank you , How am solving by activated of service program 4 , yes or not regards
March 29, 2011 07:54
hi friend code " error 14 " thank you
Limber Mejia
April 06, 2011 09:27
If jumper S1 on board D 920 ( 92 B / 7 ) is removed, the inverter can be operated without high voltage. The exposure is interrupted after 15 ms. An error message is displayed: ERR 14 kV-ACT
Adel Adam
re: polymobile 3 error E14
March 10, 2022 09:30
Hello sir I need Service manual for siemens poly mobil III. I h error 14
20,000+ imaging parts— GE, Philips, Siemens, Toshiba, Shimadzu & more. X-Ray Parts | New & Used | 90-Day Warranty | Global Shipping | Installs & Deinstalls | Ships by 6PM EST. rytexind@gmail.com | 727-557-7450| Trusted worldwide| BBB Accredited
preecha chaipet
polymobile 3 error E14
March 28, 2011 10:04
hello anybody
i have problem with sement polymobile 3 rotor okay but can't exposure with code error E14
what does it error mean ?
regards
to rate and post a comment
Alfonso Nino
Re: polymobile 3 error E14
March 29, 2011 07:42
kV-ACT
to rate and post a comment
preecha chaipet
Re: polymobile 3 error E14
March 31, 2011 08:20
Hi friend
thank you , How am solving by activated of service program 4 , yes or not
regards
to rate and post a comment
preecha chaipet
Re: polymobile 3 error E14
March 29, 2011 07:54
hi friend
code " error 14 "
thank you
to rate and post a comment
Limber Mejia
Re: polymobile 3 error E14
April 06, 2011 09:27
If jumper S1 on board D 920 ( 92 B / 7 ) is removed, the inverter can be operated without
high voltage.
The exposure is interrupted after 15 ms. An error message is displayed:
ERR 14 kV-ACT
to rate and post a comment
Adel Adam
re: polymobile 3 error E14
March 10, 2022 09:30
Hello sir
I need Service manual for siemens poly mobil III.
I h error 14
to rate and post a comment