hi i have a problem with polymobile 3 it indicates error 9,(hight pression in the tank);but on the schématic i foud that the 24v through the switch is absent ,so i go and verify the powers :when i swith on only k1 and hr are activated but not k4 and k3
does j21 opto connect the four relays ?
OR K4 and K3 are switched by the micro controller?
i m waiting for your help thanks
From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729
djamel amira
polymobile3 - error 9
September 28, 2010 05:27
hi i have a problem with polymobile 3 it indicates error 9,(hight pression in the tank);but on the schématic i foud that the 24v through the switch is absent ,so i go and verify the powers :when i swith on only k1 and hr are activated but not k4 and k3
does j21 opto connect the four relays ?
OR K4 and K3 are switched by the micro controller?
i m waiting for your help thanks
to rate and post a comment
Jose Vallejos
Re: polymobile3 - error 9
September 30, 2010 09:31
Hi Djamel,
The error 9 is mA measured 0 in idle....
if you need the service manual please send me a email to jose.vallejos@live.cl
greetings.