hi i have a problem with polymobile 3 it indicates error 9,(hight pression in the tank);but on the schématic i foud that the 24v through the switch is absent ,so i go and verify the powers :when i swith on only k1 and hr are activated but not k4 and k3
does j21 opto connect the four relays ?
OR K4 and K3 are switched by the micro controller?
i m waiting for your help thanks

Log inor Register

to rate and post a comment

(259)
(59)
(3)

Jose Vallejos

Re: polymobile3 - error 9

September 30, 2010 09:31

Hi Djamel,
The error 9 is mA measured 0 in idle....

if you need the service manual please send me a email to jose.vallejos@live.cl
greetings.