Over 150 Missouri Auctions End Today - Bid Now

PQ 2000 problem

Selcuk Akin

PQ 2000 problem

May 20, 2009 04:17

I have a problem with Picket PQ 2000 CT
Error Code:512 XSC+ANB MAX RETRIES ERROR

Case1)
· When the machine is ON,Tube rotates once or twice to reach ıts original stage and stops
· No Ready (Green) Light
Case2)
· How could we connect the laser camera to PQ 2000,(AE Tıtle, IP Addresses, Local and Remote nodes,etc…)
· When Laser and PQ connected, we get ping but no PRINT happens
· How could we see, if the Dıcom is avaibable on this Licanse· Is ‘how CNF’ a right command to see the Dıcom or any other precedure
· If Dıcom is there, we still need mım or not(digital convertor)

thank you

Selcuk akin

Log inor Register

to rate and post a comment

(18)

ORLANDO CRESPO

Re: PQ 2000 problem

May 20, 2009 05:26

HI SELCUK:
YOUR 512 PROBLEM IS DIRTY SLIP RING.
YOU CAN SELECT YOUR PRINTER WRITING IQ>CAMERA- YOU SHOULD BE ABLE TO SELECT THE ONE YOU ARE USING.
TO SEE YOUR DICOM CONFIGURATION WRITE IQ>NET_CONF OR NET_CNF, SHOW_CNF; I DONT REMEBER EXACTLY BUT ONE OF THOSE IS. GOT LUCK.

Log inor Register

to rate and post a comment

(18)

ORLANDO CRESPO

Re: PQ 2000 problem

May 21, 2009 12:12

netconfig

Log inor Register

to rate and post a comment

(33)
(4)
(8)

Alexander Tonev

Re: PQ 2000 problem

May 21, 2009 04:23

No DICOM print on PQ series.
You can only DICOM send.
If this is a new installation, for 512 error you can check the cable between console and gantry, other possible problem is operation software, bad XSC and many other.

Log inor Register

to rate and post a comment

(47)
(2)
(8)

Sergio Leonardo Barral

Re: PQ 2000 problem

May 22, 2009 12:34

Selcuk:

512 is a serial comms problem.. In your case is maximum retries: it means no comms at all, then you won't get the "ready" nor laser on. The XSC+ has minifuses on it: check them. If blown pay attention to the signal brush block: check for "hairs" shortening from one slot to other. If so, repair (only if possible...maybe not if largely worn out) and check the GPU serial comms transceivers: they could be blown too. Are the SIF port working well?. If signal brushes shortened slots (specially 120 V ones to low signal) you can have additional boards damaged. There is a special procedure if that is the case because the number of boards damaged could be higher than expected. After having the CAT scanner back to operation, we can talk about DICOM print. It works pretty fine and is very easy to set up.
Have a nice day
Sergio

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

Re: PQ 2000 problem

May 23, 2009 12:35

Hi Secul:
Sergio is right. Sometimes, little fuses are located on motherboard of the cards baskets located right side of Gantry.
You will need to clean rings too. Use 99%ethilic alcohol.
If you find a short in signal brushes, probably XSC, GPU and/or SIF are damaged. Sometimes they are reparable changing the ports but I strongly recomended to change it not repair.
You will need to change cable from SIF to Card basket motherboard too and connections between gpu and xsc (when this is broken you can do manual exposure, warm up but no scan or pilots or the shoot is maded but no x ray presented)
If you have not manuals, please tell me to look for pins and remit it to you

Best regards.

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

Re: PQ 2000 problem

May 24, 2009 07:17

Hi Selkuc:
rectifying, you don't need to change the small cable from SIF to Gantry, just test it
Best regards
Jose

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Monitor Deserves Expert Hands

Specializing in GE, Southeastern Biomedical delivers precise EKG monitor repairs and expert support you can trust. 888-310-7322.

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)