It means the system has not completed the precharge cycle, sensed by Analog Interface PCB from +200V (B+) on Generator Driver PCB.
The PRECHARGE signal passes through the Motherboard and X-ray Regulator PCB, energizing Relay K1 located in the Generator Controller Assembly. This allows charging of the pre-charge capacitors C2 & C3 from the batteries/battery charger through relay K1 contacts 6 & 9 and resistor R1. Resistor R1 limits the initial in-rush current to the
capacitors.
The voltage across the pre-charge capacitors C2 & C3 is applied through fuse F3 on the Generator Driver PCB to the voltage divider R10/R11. This voltage divider scales the precharge capacitor voltage to approximately 2.5% of the original value. This scaled voltage becomes the +200VSNS signal. The +200VSNS signal is monitored by the Technique
Processor PCB through A/D Mux U3-8 on the Analog Interface PCB.
If the Technique processor never sees the +200VSNS signal reach approximately 4.5 VDC during the boot sequence, a pre-charge error is detected.
The message PRE-CHARGE FAIL will be displayed on the Control Panel display if the Technique Processor PCB does not see the +200VSNS signal reach approximately 4.5VDC during the boot sequence (Approximately 185 VDC across pre-charge capacitors).
This is a fatal error that prevents the system boot from completing.
Hello. Most likely the problem is with the batteries. If the batteries are OK, check the charging voltage - the procedure is described in the service manual.
Lightly used system available direct from the end user customer. MONARCH QUEST
Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.
Gary Benitez
PRECHARGE FAIL ERROR CODE in an OEC 9600.
October 05, 2017 09:04
Dear Colleagues,
What does this error code mean in an OEC 9600.
PRECHARGE FAIL?
to rate and post a comment
R GOPAL
re: PRECHARGE FAIL ERROR CODE in an OEC 9600.
October 06, 2017 10:03
Either the batteries are bad or 12vdc supply is missing in Generator driver PCB.
aparnaelectromech@gmail.com
to rate and post a comment
CRAIG WAMBURA
re: PRECHARGE FAIL ERROR CODE in an OEC 9600.
October 06, 2017 10:05
It means the system has not completed the precharge cycle, sensed by Analog Interface PCB from +200V (B+) on Generator Driver PCB.
The PRECHARGE signal passes through the Motherboard and X-ray Regulator PCB, energizing Relay K1 located in the Generator Controller Assembly. This allows charging of the pre-charge capacitors C2 & C3 from the batteries/battery charger through relay K1 contacts 6 & 9 and resistor R1. Resistor R1 limits the initial in-rush current to the
capacitors.
The voltage across the pre-charge capacitors C2 & C3 is applied through fuse F3 on the Generator Driver PCB to the voltage divider R10/R11. This voltage divider scales the precharge capacitor voltage to approximately 2.5% of the original value. This scaled voltage becomes the +200VSNS signal. The +200VSNS signal is monitored by the Technique
Processor PCB through A/D Mux U3-8 on the Analog Interface PCB.
If the Technique processor never sees the +200VSNS signal reach approximately 4.5 VDC during the boot sequence, a pre-charge error is detected.
The message PRE-CHARGE FAIL will be displayed on the Control Panel display if the Technique Processor PCB does not see the +200VSNS signal reach approximately 4.5VDC during the boot sequence (Approximately 185 VDC across pre-charge capacitors).
This is a fatal error that prevents the system boot from completing.
to rate and post a comment
SupertekDave
re: re: PRECHARGE FAIL ERROR CODE in an OEC 9600.
April 29, 2018 11:56
Hi Craig,
Is the same sequence true for OEC 9800?
Thanks
to rate and post a comment
Mikhail Zakurdaew
re: PRECHARGE FAIL ERROR CODE in an OEC 9600.
October 06, 2017 10:05
Hello. Most likely the problem is with the batteries. If the batteries are OK, check the charging voltage - the procedure is described in the service manual.
to rate and post a comment
Dennis Deatrick
re: re: PRECHARGE FAIL ERROR CODE in an OEC 9600.
July 31, 2019 10:10
are the batteries available otc or do i have to get them from GE
to rate and post a comment