Precision Rxi ERR: 41

Tags: Radiology - R/F, GE

(71)
(13)

Ramil Mirshanov

Precision Rxi ERR: 41

March 14, 2016 03:14

Hello! Somebody can help to find service manual and schematics on Precision Rxi, the following message appears:Generator Offline.... my mail: ramish73@mail.ru
Thanks!

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Precision Rxi ERR: 41

March 15, 2016 08:29

yes I will zip and upload for you

1.28 Alarm 041
Code 041 refers to the transmission of the focal point and in particular to the position transducer (reference photocell).
The activation of this alarm stops the equipment, it disables the power transformer connection relay. All movements are disabled, the angle of tilt display flashes and shows the alarm code while the remaining displays switch off.
Events that may cause the error
Problem solution
Microswitch 3FC5 not functional
Replace microswitch.
Power amplification board 3S3 not functional
Replace power amplification board.
Tie cable disconnected between photocell 3FC5 and board 3S3
Reconnect cable.
Interruption in tie cable between photocell 3FC5 and board 3S3
Replace cable.
Tie cable disconnected between board 3S3 and 24S9
Reconnect cable.
CAUTION must be exercised with the screen earth connection.
Events that may cause the error
Problem solution
Interruption in tie cable between board 3S3 and 24S9
Replace cable.
Incorrect connections in cables
Refer to functional diagrams to restore connections.
Tracks interruptions on boards
Restore connections or replace boards.
Rosin joints on connectors
Restore joints or replace boards.
Interruption in flat-cable 25S3CN2-25S1CN2
Replace flat-cable.

1 person liked this
Log inor Register

to rate and post a comment

(71)
(13)

Ramil Mirshanov

re: re: Precision Rxi ERR: 41

March 16, 2016 09:10

Thank you very much for such a detailed report Mr.Place. With respect to you, Ramil

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)