We have a Prestilix 1600X room with an MPG80
generator.
On turning the machine on, we have an error 112;
this message then disappears, and the machine can be used normally.
Could anyone comment on this error ?
How could the problem can be solved ?
the battery has reached three years since install or replacement for code 501; batteries on CPU MPG1 A5 A6 and Progrm-X control bd. MPG4 A1: see Service manual vol. 1 chap. 4 to reset the date in maint. mode.
112 error on power-up B4 opto coupler of Int face bd. MPG1 A5 A11, on the MPG1 A1 A2 bd. at 19 and 20 check 5mA/50KHz and 100mA/10KHz DR job card 017 calibrate and check these points.
not sure where the 1713 is comming from ....???
Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.
Roger Vanclooster
Prestilix 1600X Error
August 16, 2007 08:25
Dear all,
We have a Prestilix 1600X room with an MPG80
generator.
On turning the machine on, we have an error 112;
this message then disappears, and the machine can be used normally.
Could anyone comment on this error ?
How could the problem can be solved ?
Many thanks,
R. Vanclooster.
to rate and post a comment
Roger Vanclooster
Re: Prestilix 1600X Error
August 16, 2007 01:59
To complete my question :
-sometimes in starting up, we have the error 501;
-in fluoroscopy we obtain the error 1713.
Any help would be deeply appreciated.
R. Vanclooster.
to rate and post a comment
Staate Hayward
Re: Prestilix 1600X Error
August 21, 2007 09:36
the battery has reached three years since install or replacement for code 501; batteries on CPU MPG1 A5 A6 and Progrm-X control bd. MPG4 A1: see Service manual vol. 1 chap. 4 to reset the date in maint. mode.
112 error on power-up B4 opto coupler of Int face bd. MPG1 A5 A11, on the MPG1 A1 A2 bd. at 19 and 20 check 5mA/50KHz and 100mA/10KHz DR job card 017 calibrate and check these points.
not sure where the 1713 is comming from ....???
to rate and post a comment
Staate Hayward
Re: Prestilix 1600X Error
August 21, 2007 09:22
mA measurement fault 112 of the generator.
to rate and post a comment
Roger Vanclooster
Re: Prestilix 1600X Error
August 22, 2007 03:33
Many thanks, Mr. Hayward, for your thoughtful answers.
Sincerely yours,
r.vanclooster@skynet.be
to rate and post a comment