Ecio Juliano

Problem for GE Vivid E 9

January 30, 2017 10:42

Hi everyone, I have a vivid E9 that presents the error HWCAP invalid DRX DC Offsets, anyone could help.

Thanks best regards.

Log inor Register

to rate and post a comment

(42)
(6)
(7)

Joseph Asmaro

re: Problem for GE Vivid E 9

January 31, 2017 10:19

Ecio,

You need to log into service mode, and reset the DC offset from there and that shall correct the issue, best of luck!

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: Problem for GE Vivid E 9

January 31, 2017 10:24

Perform DC Offset calibrations.

How is your +24V power supply? You need to check the power supply. Vacuum/blow the power supply and input filters.

1 person liked this
Log inor Register

to rate and post a comment