Problem GE Sytec on Installation

(4)

Biju Thomas

Problem GE Sytec on Installation

November 06, 2006 10:41

Note the Error Details. (Sytec s) after Installation. 1. When the System is turned ON the buttons (Table set, Ready & Start)lights and goes off with date display screen and the system gives the following errors. ER-112 XG Communication Error 01 (Description -Hard Error (Parity Error, Framing Error or Over run Error) ER-116 XG Response Timeout Then the system shows system preparation and never gets ready. Also the above error displays repeatedly in an interval. 2. Gantry seems OK as there is no error indication in the TGP Board and the Gantry passes all the off line test from the TGP Board. Also the Gantry rotates with power ON and comes back to the home position. 3. The Generator side is OK as there is exposure when it is made locally from the generator. Whether the buttons (Table set, Ready & Start) lights when the system is turned ON. This lights even with the CPW board removed or with all connection between Gantry, Generator with MISC is removed (2 flat cables). Cable continuety is good. Is this due to any safety loop. Hope somebody can assist me. With Best Regards, Biju

Log inor Register

to rate and post a comment

Umberto Sciunnach

Re: Problem GE Sytec on Installation

November 06, 2006 01:35

OK
You have installation problem . Let me know more details because I can help you but I stay in Italy.
let me know.regards
uto

Log inor Register

to rate and post a comment

(2)

Suresh Kumar thulukkanam

Re: Problem GE Sytec on Installation

November 07, 2006 10:41

Dear Biju,
It seems to be problem in the Console ..Can you check the OFF-line test in the CPW board by selecting proper DIP switch setting in the CPW board as well as the communication between Hard disk and CPW board and other associated board functional intergrity in the console Okay? .

If you specify inform me the Sytec Model?I can assist you further more about trouble shooting .

Regards,

Suresh

Log inor Register

to rate and post a comment

daniel zhao

Re: Problem GE Sytec on Installation

November 07, 2006 11:23

Dear Biju,

Can you check the 5v PS in the console? maybe this PS is fault.

Log inor Register

to rate and post a comment

(4)

Biju Thomas

Re: Problem GE Sytec on Installation

November 07, 2006 11:41


Thanks Suresh,

The Model is Sytec S.

I did both the offline test and it passes.

Also if we make the DIP 4 of SW6A27 (OC INTF SW) of MISC Board ON the system starts. By setting the DIP-4 ON the communication between Console and TGP is disabled. So the console is OK except MISC Board.

In the other case after the Date & Time display the system remains in preparation.

This System when turned ON the buttons (Table set, Ready & Start) lights, is it normal.

Regards,

Biju

Log inor Register

to rate and post a comment

(16)
(1)

larry zhuang

Re: Problem GE Sytec on Installation

November 08, 2006 11:09

Dear Mr.Biju:
I Think it's indeed the communication problem as the sysytem shows,You could check
1.the +5VDC /120A power in OC,adjust it to the specification,2.and all the DC power in XG cabinet.3.the interface ICs on the two B'd :
power contral and the i/f b'd in the OC.
when you p-on the system ,some of the putton light on OC could be lit up.---it's OK.

Log inor Register

to rate and post a comment

(72)
(1)

Elliot Marcus

Re: Problem GE Sytec on Installation

November 08, 2006 11:09

Biju,

Since you say there are no errors led's illuminated on the TGP board, the Gantry performs its initial powerup rotate-rotate sequence and no other errors other than those in the log, in my opinion you have either a bad cable from the Misc board to the Generator (Misc CN7 to XG Power Cont Assy CN1) or either a bad Misc or Power Cont Board.

Hope this helps,

Elliot

Log inor Register

to rate and post a comment

(4)

Biju Thomas

Re: Problem GE Sytec on Installation

November 09, 2006 12:52


Hai All,

Thanks

Console 5v/120A power is OK and the continuity of the cable from MISC to TGP & Power Control board is OK.

Regards,

Biju

Log inor Register

to rate and post a comment

Dean Kenney

Re: Problem GE Sytec on Installation

November 09, 2006 10:25

their are sytec i series (slj), sytec 2000/3000/4000/vertex series (Vj)(Lj),,sytec synergy series (XM1500),, sytec 1600i/1800i/Eu Series (SYj),,sytec SRi/Legato Series (XMj),, So which one is it specificly it might as simple as a corroded connection or you need to jumper out an interlock. I know systems that will do everything except an exposure when a door is not closed.

Log inor Register

to rate and post a comment

(4)

Biju Thomas

Re: Problem GE Sytec on Installation

November 11, 2006 02:26


Hai Dean/All,

Thanks.

The Model is Sytec S (2000i). This belongs to Sytec i Series(SLj)

I also suspect some interlock problem.

Awaiting your reply.

Best Regards,

Biju


Log inor Register

to rate and post a comment

(2)

SURESH KUMAR

Re: Problem GE Sytec on Installation

November 14, 2006 10:02

Dear BIJU ,

Probablly please check gantry front & side covers are properly locked. Also check for cable take up switches at rear side of gantry.

with regards
Suresh

Log inor Register

to rate and post a comment

(72)
(1)

Elliot Marcus

Re: Problem GE Sytec on Installation

November 14, 2006 03:50

Suresh,

Gantry Interlocks and/or a Cable Takeup switch would not cause the ER-112 and ER-116 errors Biju spoke of initially in this thread. Since Biju has checked the cables from the Misc to TGP and Pwr Cont boards, the problem is most likely a defective Misc or Power Control board.

Log inor Register

to rate and post a comment

(120)
(6)
(1)

manuel valdez

re: Re: Problem GE Sytec on Installation

March 14, 2013 06:12

hi Elliot i have a problem with a sytec 2000i could you helpme please?
1. I found a purpose general diode in the axial motor in short circuit
2 this diode burn the 24 vd power supply
3. I changed the diode
4. i repaired the power supply
but now i turn on the system and inmediatly the servo-amplifier show the error 8
and the tTGP board light on the led SVALM I test the servo-amplifier in another system and it work perfectly same the TGP boar, I changed the motor and I cant fix teh problem
any help will be apreciated

Log inor Register

to rate and post a comment