I give maintenance to a Gyroscan 5NT and this show low pressure inside the magneto. It show 20 mbar all time I try to make rise it using the MAIN HEATER but don“t function. The pressure gauge work perfect.

Log inor Register

to rate and post a comment

(42)
(1)
(17)

Marc Fessler

Re: Problem, Gyronscan 5NT without pressure

March 16, 2009 02:29

Please check that the Helium fill port sealing nut is tight and that there is an o-ring there as well.
You should also check that the burst disc is not leaking, also check the sealing rings and o-rings on the ramp lead ports. In other words you're looking for a leak.

Log inor Register

to rate and post a comment

(57)

(1)

Bagher Mahdavi

Re: Problem, Gyronscan 5NT without pressure

March 16, 2009 02:47

Hi
try to check the burst disk.
if it is OK and helium consumption is not to much then somewhere is leaking try to find it with foam.

Best
Bagher

Log inor Register

to rate and post a comment