Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Problem in a LORAD IV Hard disk drive

Tags: Mammo

(45)
(6)
(1)

Jorge Flores

Problem in a LORAD IV Hard disk drive

December 23, 2014 10:20

Dear friends, I have a problem on a LORAD IV. At turn on gives the message "NON SYSTEM DISK OR DISK ERROR. REPALCE AND STRIKE ANY KEY". This is the frist time I found a LORAD IV system with a HDD!!I need to replace that disk. I don´t have any info about it and how to load the system software when the disk is replace. There is not a floopy disk with the SFW in the site.

Thanks in advance for your help.

2 people liked this
Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Problem in a LORAD IV Hard disk drive

December 24, 2014 01:31

the drive is located in the operator console. have you tried reseating the cable and verify the power supplies are good? I never reloaded but I have cloned another drive then redid the configurations.
but if you have the floppy insert in the console floppy drive have power off the turn unit on it will load the software from the floppy to the harddrive. just select language. it will then show remove floppy and reboot, it will take longer than normal to reboot due to unit rebuilding patient database.
I will see about getting a copy of the floppy just would need to know version you need

Log inor Register

to rate and post a comment

(45)
(6)
(1)

Jorge Flores

re: re: Problem in a LORAD IV Hard disk drive

December 24, 2014 10:02

Thanks for your tips. How can I know the versión?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: Problem in a LORAD IV Hard disk drive

December 24, 2014 01:37

let me know what eeprom version you have on the CPU board in the console and in the generator

Log inor Register

to rate and post a comment

(45)
(6)
(1)

Jorge Flores

re: re: re: re: Problem in a LORAD IV Hard disk drive

December 28, 2014 01:13

Lenny...the site is out of town and nobody can verify that. At turn on this is what You can see in the screen " CARD-386 bios versión 3.10C"
Does that help?
Best regards

Log inor Register

to rate and post a comment

(45)
(6)
(1)

Jorge Flores

re: re: re: re: Problem in a LORAD IV Hard disk drive

December 28, 2014 01:16

Lenny...the site is out of town and no one can check that. At turn on on the screen You can see this "Card-386 bios versión 3.10C"...Does that helps?
Best regards

Log inor Register

to rate and post a comment

(45)
(6)
(1)

Jorge Flores

re: re: re: re: Problem in a LORAD IV Hard disk drive

December 28, 2014 01:16

Lenny: The site is out of town and there in anyone that can check it but at turn on on the screen you can see this "CARD-386 BIOS Version 3.10C" Do that help?
Best regards,

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: re: Problem in a LORAD IV Hard disk drive

December 29, 2014 11:40

no not really that is just the BIOS version. I am trying to get a few different copy of the disks from a friend however he is on vacation until next week

Log inor Register

to rate and post a comment

(45)
(6)
(1)

Jorge Flores

re: re: re: re: re: re: Problem in a LORAD IV Hard disk drive

December 30, 2014 10:11

Sounds good...please let me know...The best for the 2015

Log inor Register

to rate and post a comment

(15)

LUIS CASTRO

re: re: Problem in a LORAD IV Hard disk drive

June 01, 2015 04:05

Hi Lenny,
I Have hdd problems my system still starts but sometimes stays frozen, the operator console is ver. 2.1.1, do you have the floppy disk?

Luis Castro

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: Problem in a LORAD IV Hard disk drive

June 01, 2015 05:21

yes sending it via sendthisfile

Log inor Register

to rate and post a comment

(28)
(1)

Larry Underriner

re: Problem in a LORAD IV Hard disk drive

December 24, 2014 01:31

Try the guys at Digitec Medical. They refurbish MIV's and also offer Tech Support and Parts also. Highly recommend them.

Log inor Register

to rate and post a comment

(45)
(6)
(1)

Jorge Flores

re: re: Problem in a LORAD IV Hard disk drive

December 24, 2014 10:02

Thanks I will

Log inor Register

to rate and post a comment

(5)

MAMA

re: Problem in a LORAD IV Hard disk drive

December 24, 2014 01:32

i have not seen HDD with this unit ??

Log inor Register

to rate and post a comment

(5)

MAMA

re: Problem in a LORAD IV Hard disk drive

December 24, 2014 01:32

pls mail me your mail ID

Log inor Register

to rate and post a comment

(45)
(6)
(1)

Jorge Flores

re: re: Problem in a LORAD IV Hard disk drive

December 24, 2014 10:02

jcfg_99@yahoo.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)