problem in a Sytec Sri CT

May 30, 2008 09:08

I need some techinical help for GE CAT Scan,I have a problem in a Sytec Sri CT, the errors : Scan ERROR 101 CODE 000053 followed by XG ERROR:117 CODE :483.Can anybody help me in this problem?

Regards

Log inor Register

to rate and post a comment

(43)

(2)

Ernesto Guerra

Re: problem in a Sytec Sri CT

May 30, 2008 11:31

Mr. Mustafa check your axial drive, is the system rotating ok? If it is then you 117 is either on your x-ray tube, high tank. Let me know. thanks

Log inor Register

to rate and post a comment

(57)
(3)
(6)

R.K.S singh

Re: problem in a Sytec Sri CT

May 31, 2008 04:19

DEAR
I CAN HELP U PLEASE GIVE ME ERROR CODE AND SUB CODE LIKE U HAVE ER 117 XG ERROR WHAT IS SUB CODE AND WHAT HAPPEN PLEASE TELL IN BREIF I CAN EXPLIANE U WELL
THANKS & REGADS
ROHIT SINGH

Log inor Register

to rate and post a comment

(74)
(5)
(10)

Tayfun Canli

Re: problem in a Sytec Sri CT

June 01, 2008 05:48

Hi

Please let me know the detail of your problem;

1) Perform an offline scan with XG off, Rotor Off; Note error codes
2) Perform an offline scan with XG off, Rotor ON; Note error codes
3) Perform an offline scan with XG ON (min KV, min MA and min time), Rotor ON ; Note error codes.

My Best Wishes
Tayfun CANLI
Link Medical Systems

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Source for Sony Medical Displays, Printers & More!

Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)