Problem in GE AMX4 Plus, 45E Generator Error

Tags: Radiology - R/F, GE HEALTHCARE

(83)
(18)

GH

Problem in GE AMX4 Plus, 45E Generator Error

September 14, 2022 11:06

Dear All,
I have problem in GE AMX4 Plus, it gives error 45E.
Can any one help me in that.
Is there any suspected other than the filament board or HV cables?
Regards,
Ghassan

Log inor Register

to rate and post a comment

(10)
(1)

Okuwin

re: Problem in GE AMX4 Plus, 45E Generator Error

September 15, 2022 09:56

Try these fixes.
ERROR 45E
1. Check For 2 KHZ Phase 1 & 2 Signals At J2-17 & 19 Of FIL/KVP Board
2. Check For Continuity Between J5 Plug On FIL/KVP Board And Filament Terminals On HV Xformer
3. Check For Open Or Shorted Tube Small Filament – Is The Small focal Spot Lit During Prep?
A 6V Battery Voltage Can Be Applied Directly To The Filament To See If It Will Lite --- Even Though A Continuity Check Of The Filament Might Indicate It’s Ok
4. Check For Defective HV Cathode Cable (Internal Short)
5. Replace FIL/KVP Board - Check The +24 Volts To J1-2 Of FIL/KVP Board
6. Is The XRT From Another Vendor? (Must Have Internal Resistor In Series With Filament)
7. Bad High Voltage Divider (H/V Bleeder)
8. FIL SHORTED (Check J2-25 On FIL/KVP Board)

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. πŸ“ž 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)