PROBLEM IN. GE. LIGHTSPEED VCT 64.

January 08, 2018 09:33


Dear Sirs.

I need help from you.

I'm repairing TAC, brand: GE, Model: Lightspeed VCT 64, software version GOC6,5.

Description of the problem:
1.- Allows the heating of the x-ray tube.
2.-It does not allow fast tube calibration and this message appears in log error:
Host:tgp Proc: tgc_subsys
File: scTrigSeqActions.c Line: 584
Host: Table Gantry Processor. Ermes# : 260100152
Exeption Class: Abort Severity: Primary
File: scTrigSeqActions.c Line:584
Function : System Monitoring : Xray Exposure Interlocks
Scan Type : Axial Scan: 50006/2 /3
Exposure Interlock Opened.
The current loop from Gantry to PDU.
The Cause of the failure: Console Relay.

Could you light the way .............

Thank you very much for your time and effort.

My email is: jose.ig.herrera@gmail.com

A greeting..

José

Log inor Register

to rate and post a comment

(2)

Hector Quintana

re: PROBLEM IN. GE. LIGHTSPEED VCT 64.

May 29, 2026 03:37

Hi, im having the same problem, already checked stop sensors and interconnections, what was the problem in your case?

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

re: re: PROBLEM IN. GE. LIGHTSPEED VCT 64.

June 02, 2026 10:11

Check al interlocks, include doors ones

Log inor Register