Description of the problem:
1.- Allows the heating of the x-ray tube.
2.-It does not allow fast tube calibration and this message appears in log error:
Host:tgp Proc: tgc_subsys
File: scTrigSeqActions.c Line: 584
Host: Table Gantry Processor. Ermes# : 260100152
Exeption Class: Abort Severity: Primary
File: scTrigSeqActions.c Line:584
Function : System Monitoring : Xray Exposure Interlocks
Scan Type : Axial Scan: 50006/2 /3
Exposure Interlock Opened.
The current loop from Gantry to PDU.
The Cause of the failure: Console Relay.
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
JOSE HERRERA
PROBLEM IN. GE. LIGHTSPEED VCT 64.
January 08, 2018 09:33
Dear Sirs.
I need help from you.
I'm repairing TAC, brand: GE, Model: Lightspeed VCT 64, software version GOC6,5.
Description of the problem:
1.- Allows the heating of the x-ray tube.
2.-It does not allow fast tube calibration and this message appears in log error:
Host:tgp Proc: tgc_subsys
File: scTrigSeqActions.c Line: 584
Host: Table Gantry Processor. Ermes# : 260100152
Exeption Class: Abort Severity: Primary
File: scTrigSeqActions.c Line:584
Function : System Monitoring : Xray Exposure Interlocks
Scan Type : Axial Scan: 50006/2 /3
Exposure Interlock Opened.
The current loop from Gantry to PDU.
The Cause of the failure: Console Relay.
Could you light the way .............
Thank you very much for your time and effort.
My email is: jose.ig.herrera@gmail.com
A greeting..
José
to rate and post a comment
Hector Quintana
re: PROBLEM IN. GE. LIGHTSPEED VCT 64.
May 29, 2026 03:37
Hi, im having the same problem, already checked stop sensors and interconnections, what was the problem in your case?
to rate and post a comment
Jose Morillo
re: re: PROBLEM IN. GE. LIGHTSPEED VCT 64.
June 02, 2026 10:11
Check al interlocks, include doors ones
to rate and post a comment