PROBLEM IN. GE. LIGHTSPEED VCT 64.

January 08, 2018 09:33


Dear Sirs.

I need help from you.

I'm repairing TAC, brand: GE, Model: Lightspeed VCT 64, software version GOC6,5.

Description of the problem:
1.- Allows the heating of the x-ray tube.
2.-It does not allow fast tube calibration and this message appears in log error:
Host:tgp Proc: tgc_subsys
File: scTrigSeqActions.c Line: 584
Host: Table Gantry Processor. Ermes# : 260100152
Exeption Class: Abort Severity: Primary
File: scTrigSeqActions.c Line:584
Function : System Monitoring : Xray Exposure Interlocks
Scan Type : Axial Scan: 50006/2 /3
Exposure Interlock Opened.
The current loop from Gantry to PDU.
The Cause of the failure: Console Relay.

Could you light the way .............

Thank you very much for your time and effort.

My email is: jose.ig.herrera@gmail.com

A greeting..

José

Log inor Register

to rate and post a comment

(2)

Hector Quintana

re: PROBLEM IN. GE. LIGHTSPEED VCT 64.

May 29, 2026 03:37

Hi, im having the same problem, already checked stop sensors and interconnections, what was the problem in your case?

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

re: re: PROBLEM IN. GE. LIGHTSPEED VCT 64.

June 02, 2026 10:11

Check al interlocks, include doors ones

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Therapeutic Technologies Surplus Auctions June 15th

Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)