hy every body
i have a sytem PQ5000, i facing a problem during calibration that when i saving the position of Centring Rod
the message apear that
"&& terminated center(737) with signal"
sigmentation violation-core dumped
error code(163)
i will be very thankfull for any Good Suggestion.
musain77@yahoo.com
1.- make sure you have diagnostics full available
cddiag
rev
2.-make sure pin is not on image, just rod(plastic without metal pin
3.- artifacts on images? computer will can not calculate center
Dear jose
no artifacts are coming in image,
but in last 2, 3 scans of Axial( of Brain) the Two mid lines are coming in images,
is the leteral Centring or tube Centring is must.
Your problem is a software related one. Core dumped means that regulus couldn't progress executing the task and "dumped" a core file. Dyring the "old good days" you could archive it using arcdump and send it to Picker International Service Programs for evaluation. Today, I don't know if such a help is available at Philips.
Being a software problem, start trying soft-related solutions (i.e. reinstall it)
Center can perfectly handle phantom misplacements like pin or rod position giving explicit error messages, aborting the program and returning the IQ prompt without core dumping.
I hope the above lines can give you a hint about the solution of your problem.
musain khan
problem in Picker PQ 5000
August 20, 2009 06:37
hy every body
i have a sytem PQ5000, i facing a problem during calibration that when i saving the position of Centring Rod
the message apear that
"&& terminated center(737) with signal"
sigmentation violation-core dumped
error code(163)
i will be very thankfull for any Good Suggestion.
musain77@yahoo.com
to rate and post a comment
Kadir Kuran
Re: problem in Picker PQ 5000
August 20, 2009 07:17
Did you check your compensator may be it is broke inside
to rate and post a comment
musain khan
Re: problem in Picker PQ 5000
August 22, 2009 02:05
Yes mr kadir
i already check the compensator, But the Compesnsator is OK,
IS you faced This type of Problem
to rate and post a comment
Alvaro Moreira
Re: problem in Picker PQ 5000
August 23, 2009 05:37
It looks like you're scanning the internal metal pin of the central rod. You need to make sure only acrilyc gets on the way.
Alvaro Moreira
to rate and post a comment
Jose Morillo
Re: problem in Picker PQ 5000
August 22, 2009 02:05
1.- make sure you have diagnostics full available
cddiag
rev
2.-make sure pin is not on image, just rod(plastic without metal pin
3.- artifacts on images? computer will can not calculate center
best regards and let me know
Jose
to rate and post a comment
musain khan
Re: problem in Picker PQ 5000
August 24, 2009 09:31
Dear jose
no artifacts are coming in image,
but in last 2, 3 scans of Axial( of Brain) the Two mid lines are coming in images,
is the leteral Centring or tube Centring is must.
to rate and post a comment
Sergio Leonardo Barral
Re: problem in Picker PQ 5000
August 28, 2009 11:35
Musain:
Your problem is a software related one. Core dumped means that regulus couldn't progress executing the task and "dumped" a core file. Dyring the "old good days" you could archive it using arcdump and send it to Picker International Service Programs for evaluation. Today, I don't know if such a help is available at Philips.
Being a software problem, start trying soft-related solutions (i.e. reinstall it)
Center can perfectly handle phantom misplacements like pin or rod position giving explicit error messages, aborting the program and returning the IQ prompt without core dumping.
I hope the above lines can give you a hint about the solution of your problem.
Best regards
Sergio