Most likely your acquisition interface board is the problem, since it does not show up on whole body scans but on statics. Does the lines show up on flood images? Are they sharply defined (digital) or somewhat blurred?
The acquisition interface is in the CAP. However, since the lines are blurred, it may be in the Multi head divider pcb, which is also in the CAP. Do you have diagnostics installed that you can run?
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
haidar jaafar
problem in prism 1500xp
April 05, 2013 10:11
horizantals lines appears on static study(doesnt appears) with whole body).anybody help with 10x
to rate and post a comment
William Carroll
re: problem in prism 1500xp
April 05, 2013 05:38
Most likely your acquisition interface board is the problem, since it does not show up on whole body scans but on statics. Does the lines show up on flood images? Are they sharply defined (digital) or somewhat blurred?
to rate and post a comment
haidar jaafar
re: re: problem in prism 1500xp
April 06, 2013 01:12
10x william.is the acquisation interface in the cap or that attached on the mother board of the head? the lines are blurred.
to rate and post a comment
William Carroll
re: re: re: problem in prism 1500xp
April 07, 2013 08:12
The acquisition interface is in the CAP. However, since the lines are blurred, it may be in the Multi head divider pcb, which is also in the CAP. Do you have diagnostics installed that you can run?
to rate and post a comment
haidar jaafar
re: re: re: re: problem in prism 1500xp
April 08, 2013 09:44
multi head divider you mean digital divider? yes have diagnostics but need the activation key.
to rate and post a comment
William Carroll
re: re: re: re: re: problem in prism 1500xp
April 08, 2013 12:25
Does the digital divider has inputs for one head or 3 heads?
to rate and post a comment
William Carroll
re: re: re: re: re: problem in prism 1500xp
April 08, 2013 03:43
Please email me directly and i can help you diagnose the problem.