Problem on/off Sytec 3000

October 28, 2009 11:04

Turn on the o.c. all ok!, The problem I have to turn it off from the o.c. ( not off ).He I changed all the relays in power module, and nothing happens.
Upon activating the key on / off you can hear the power of a relay in the power module.

Log inor Register

to rate and post a comment

(100)
(9)
(4)

JORGE LUIS LOPEZ

Re: Problem on/off Sytec 3000

October 28, 2009 12:29


Dino,Most likely, the push button is jammed in the on position, let me know i'm in skype: jorgellopez.
hope this helps. Good luck

Log inor Register

to rate and post a comment

(24)
(1)

dino mantegazza

Re: Problem on/off Sytec 3000

October 29, 2009 11:43

ok thanks I will check