problem Optimus 65 Philips

Tags: Radiology - R/F

(25)

(2)

Raul Cervantes

problem Optimus 65 Philips

October 31, 2014 08:57

I have a problem with Optimus 65 Philips, when power up on unit all leds on control panel are glow up, not flashing, (generator side, all boards),power supply are OK, battery on board 139 and 119 are OK (3V). display on tube side is OK.(power up OK, only system led is out)
Can any one help to solve the problema?. thanks.

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: problem Optimus 65 Philips

November 01, 2014 02:07

What do you mean system led? Does the Control desk shows the KV mas and MS?

Log inor Register

to rate and post a comment

(25)

(2)

Raul Cervantes

re: re: problem Optimus 65 Philips

November 03, 2014 08:24

Hi, Juan, the control desk, not show KV, mas and ms, only in the cabinet show the 4 leds are lit,(CU, MA, KV and interfase board) not flashing.

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: re: re: problem Optimus 65 Philips

November 03, 2014 03:40

Are all the bottoms in the control desk on? do they come on and then they go back off?

Log inor Register

to rate and post a comment

(25)

(2)

Raul Cervantes

re: re: re: re: problem Optimus 65 Philips

November 03, 2014 06:47

When the problem is present no, in few occasion the power up is ok, but in one or two minutes the display the KV, MAS and Ms, is off only lit the leds, no errors

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: re: re: re: re: problem Optimus 65 Philips

November 04, 2014 09:01

You might have a problem with the console, check Power Supplies on the console. The console boots up and then it communicates with the CPU, so if the console is not booting there are no errors.The normal boot is led's on , led's off(power on self test)and then it display kv mas and ms when it communicates

Log inor Register

to rate and post a comment

(62)
(20)
(8)

Muhammad Muhammad

re: problem Optimus 65 Philips

November 01, 2014 02:08

Hello
Kindly send us email with details your x-ray control panel led is off.no any display any error.when you power up control panel is booting or not also send us photo.
Thanks

Log inor Register

to rate and post a comment

(13)
(3)

Adnan

re: problem Optimus 65 Philips

November 01, 2014 02:10

You need a terminal to retrieve a error log by that log you will find your exact problem. If you have terminal than restore generator backup file.

Log inor Register

to rate and post a comment

(66)
(8)

Haroon Rashid

re: problem Optimus 65 Philips

November 01, 2014 02:11

You need a terminal to retrieve a error log by that log you will find your exact problem. If you have terminal than restore generator backup file.

Log inor Register

to rate and post a comment

(25)

(2)

Raul Cervantes

re: re: problem Optimus 65 Philips

November 03, 2014 08:25

Hi, Haroon, I have service manual and terminal with parallel port and Xscope software, and backup disk, but not the hardware key, can I make restore??...thanks

Log inor Register

to rate and post a comment

(66)
(8)

Haroon Rashid

re: re: re: problem Optimus 65 Philips

November 04, 2014 09:01

No you cannot restore generator without key.for further detail contact me at gems.haroon@hotmail.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)