PROBLEM REGUIS 190 CS-2
June 12, 2015 07:39
HI ,I HAVE A PROBLEM WITH REGUIS 190.
MESSAGE ON DISPLAY "IMPROPER INSERTION PRESS EJECT"
MESSAGE ON SCREEN "170-01 :IMAGE READING ERROR"
THANKS FOR YOU HELP.
to rate and post a comment
Ed Bican
re: PROBLEM REGUIS 190 CS-2
June 15, 2015 04:21
There are two infrared emitters and detectors near the barcode reader where you insert the cassette. Make sure they are clean. You can use a cotton swab and try and clean them with that or open up the assy. Make sure you clean both the emitter and detector. I always clean these during a PM.
to rate and post a comment
212665090440
re: re: PROBLEM REGUIS 190 CS-2
June 16, 2015 09:30
HI Ed Bican ;
THANKS FOR YOUR HELP. I DO THAT BUT THE PROBLEM STILL.NOW I HAVE THIS MESSAGE "ERROR 12000 RENOUVE REMAINING CST"
BEST REGARDS.
to rate and post a comment
Ed Bican
re: re: re: PROBLEM REGUIS 190 CS-2
June 16, 2015 04:15
The error indicates that there is a cassette inserted. Check all your sensors they may be dirty or bad maybe check and reseat cables to sensors. You may have to open the unit to clean the infrared sensors and emitters good.