maybe you have the display or another board with poor connection. try readjust it
what do other people think about it: if something is working well, don't touch it!!
hi ! will you check in the power supply the 5V this voltage is for echoes of transducer .if this voltage raise that 5.3 o more you will regulate to 5.0 v or minor observe the image...
Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: ryan.swiderski@cyclicmaterials.earth or call 437-826-1621
ismael ceballos
problem whit u.s. medison sa-600
March 30, 2005 10:06
i have a problem whit a u.s. medison sa-600.
after the preventive service, the image have artefact, but not all time, somebody can help me
to rate and post a comment
gustavo fiochi
Re: problem whit u.s. medison sa-600
March 30, 2005 02:54
maybe you have the display or another board with poor connection. try readjust it
what do other people think about it: if something is working well, don't touch it!!
to rate and post a comment
Jim Carr
Re: problem whit u.s. medison sa-600
March 30, 2005 04:15
Sounds like it may be a problem with a probe. Is it with all transducers or only one?
to rate and post a comment
ismael ceballos
Re: problem whit u.s. medison sa-600
March 31, 2005 04:01
they have only one probe but this artefact no see all time, i think the probe is good, what do you think,
sorry but my english not is good
to rate and post a comment
Miguel Chavira
Re: problem whit u.s. medison sa-600
April 01, 2005 12:38
hi ! will you check in the power supply the 5V this voltage is for echoes of transducer .if this voltage raise that 5.3 o more you will regulate to 5.0 v or minor observe the image...
to rate and post a comment
ismael ceballos
Re: problem whit u.s. medison sa-600
April 06, 2005 10:07
thanks to all, now is work well, was a poor contact.
to rate and post a comment