FRIENDS,
I have a problem with a Bennett Contour Plus.
The problem started with the downward compression. When I pressed the pedal, the compressor, goes down, but it doesnt stop when I release the pedal. After that, it stopped working the downward compression, and it doesnt communicate with the control module. Now, When I press the pedal the Light turns on, the relay activate, but the compressor does not go down.
Please, let me know if someone can help, or provide the technical manual.
Thanks in advance.
Eduardo Firmiano
GBS offers service, repair, rentals & spare parts for your ventilators. With over 40 years of experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449
Eduardo Firmiano
PROBLEM WITH A BENNETT CONTOUR PLUS
November 06, 2006 01:29
FRIENDS, I have a problem with a Bennett Contour Plus. The problem started with the downward compression. When I pressed the pedal, the compressor, goes down, but it doesnt stop when I release the pedal. After that, it stopped working the downward compression, and it doesnt communicate with the control module. Now, When I press the pedal the Light turns on, the relay activate, but the compressor does not go down. Please, let me know if someone can help, or provide the technical manual. Thanks in advance. Eduardo Firmiano
to rate and post a comment
Howard Karp
Re: PROBLEM WITH A BENNETT CONTOUR PLUS
November 14, 2006 09:41
check the connections on the compression drive motor. I had a similar problem and found a cold solder connection at the motor under the heat shrink.