PROBLEM WITH A BENNETT CONTOUR PLUS

November 06, 2006 01:29

FRIENDS, I have a problem with a Bennett Contour Plus. The problem started with the downward compression. When I pressed the pedal, the compressor, goes down, but it doesnt stop when I release the pedal. After that, it stopped working the downward compression, and it doesnt communicate with the control module. Now, When I press the pedal the Light turns on, the relay activate, but the compressor does not go down. Please, let me know if someone can help, or provide the technical manual. Thanks in advance. Eduardo Firmiano

Log inor Register

to rate and post a comment

(28)
(1)
(1)

Howard Karp

Re: PROBLEM WITH A BENNETT CONTOUR PLUS

November 14, 2006 09:41

check the connections on the compression drive motor. I had a similar problem and found a cold solder connection at the motor under the heat shrink.