Problem with a GE ADVANTX MPPU R/F SYSTEM.

Tags: Radiology - R/F, GE

(123)
(12)
(7)

Manuel Brugat

Problem with a GE ADVANTX MPPU R/F SYSTEM.

May 08, 2017 10:09

Dear Colleagues:
I have this problem with a GE Advantx with MPPU generator. The screen on the console does not light up, the generator does not turn on and the table does not power-up either.
I have troubleshoot as much as I could but everything indicates the source of the problem comes from the logic circuits of the positioner.

Can anyone give me a hint?

Sincerely:

Manuel Brugat/Image Instrumentation, Inc.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Problem with a GE ADVANTX MPPU R/F SYSTEM.

May 09, 2017 10:11

The console should still come on even if there is a positioner error.
Is the neon NE1 illuminated on the MPPU2A1A4?
If yes check the 24vdc on this board.