Dear Colleagues:
I have this problem with a GE Advantx with MPPU generator. The screen on the console does not light up, the generator does not turn on and the table does not power-up either.
I have troubleshoot as much as I could but everything indicates the source of the problem comes from the logic circuits of the positioner.
The console should still come on even if there is a positioner error.
Is the neon NE1 illuminated on the MPPU2A1A4?
If yes check the 24vdc on this board.
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
Manuel Brugat
Problem with a GE ADVANTX MPPU R/F SYSTEM.
May 08, 2017 10:09
Dear Colleagues:
I have this problem with a GE Advantx with MPPU generator. The screen on the console does not light up, the generator does not turn on and the table does not power-up either.
I have troubleshoot as much as I could but everything indicates the source of the problem comes from the logic circuits of the positioner.
Can anyone give me a hint?
Sincerely:
Manuel Brugat/Image Instrumentation, Inc.
to rate and post a comment
Lenny Place
re: Problem with a GE ADVANTX MPPU R/F SYSTEM.
May 09, 2017 10:11
The console should still come on even if there is a positioner error.
Is the neon NE1 illuminated on the MPPU2A1A4?
If yes check the 24vdc on this board.
to rate and post a comment