I JUST USED THE SYSTEM TWICE AND NOW IT DOEST WANT TO LOAD IT HANGS IN THE FIRST PART OF THE SOFTWARE LOADING AND STOPS THERE.
I GOT NEW PROBES DOES THIS CAN CAUSE THIS? DO YOU THINK I HAVE TO RELOAD THE SOFTWARE?
ILL APRECITE ANY COMENT.
HOW CAN I FIND OUT WHAT'S THE PROBLEM
Rossana,
It is possible although not too likely that one of the probes is keeping the system from booting. It should boot-up without any probes attached, so try that and see what happens. If it still doesn't boot, try running diagnostics. Are you getting an error message?
Note that the 400CL model does not run any sector probes.
You can download the service manual by going to http://apps.gehealthcare.com/servlet/ClientServlet?REQ=RNEW&MODALITY=US
Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.
rossana diaz
PROBLEM WITH A LOGIQ 400CL PRO
June 09, 2007 08:48
I JUST USED THE SYSTEM TWICE AND NOW IT DOEST WANT TO LOAD IT HANGS IN THE FIRST PART OF THE SOFTWARE LOADING AND STOPS THERE.
I GOT NEW PROBES DOES THIS CAN CAUSE THIS? DO YOU THINK I HAVE TO RELOAD THE SOFTWARE?
ILL APRECITE ANY COMENT.
HOW CAN I FIND OUT WHAT'S THE PROBLEM
to rate and post a comment
Jim Carr
Re: PROBLEM WITH A LOGIQ 400CL PRO
June 11, 2007 11:04
Rossana,
It is possible although not too likely that one of the probes is keeping the system from booting. It should boot-up without any probes attached, so try that and see what happens. If it still doesn't boot, try running diagnostics. Are you getting an error message?
Note that the 400CL model does not run any sector probes.
You can download the service manual by going to http://apps.gehealthcare.com/servlet/ClientServlet?REQ=RNEW&MODALITY=US
to rate and post a comment