Hai,
i have a problem with AMX4,please help me to sort out.
1. Intermittent exposure problem. I replaced the handswitch and also checked all contactors when press the exposure handswitch. All the relevent contactors are working when press the handswitch and after long delay immediately it expose. Any idea?
2. Drive problem: Driving as normal condition when it drive and suddenly starts slow and making to push. I checked the battery and drive board. It seems Ok.. Any advice?
With the drive problem, you're probably having a problem with the tube park switch. Where the tube locks down there is a switch that probably needs adjusted up or replaced.
The intermittent exposure problem can still be a contactor .Did you test contactors in diagnostic mode ? As far as your drive problem , you probabaly have aproblem with the tube either not sitting properly in cradle or the the sensor .If you have any problems with unit you can email your questions directly ,I will try to help out.
rlatkovich@alpha-imaging.com
One call to 440-263-6575 and your move will be covered.
We repair & store trailers. We build to your custom needs.
JDI Mobile Medical Manufacturing. Inc. We have your back.
Thameen Ansari
Problem with AMX 4
June 20, 2008 05:14
Hai,
i have a problem with AMX4,please help me to sort out.
1. Intermittent exposure problem. I replaced the handswitch and also checked all contactors when press the exposure handswitch. All the relevent contactors are working when press the handswitch and after long delay immediately it expose. Any idea?
2. Drive problem: Driving as normal condition when it drive and suddenly starts slow and making to push. I checked the battery and drive board. It seems Ok.. Any advice?
to rate and post a comment
Gary Fink
Re: Problem with AMX 4
June 20, 2008 03:07
With the drive problem, you're probably having a problem with the tube park switch. Where the tube locks down there is a switch that probably needs adjusted up or replaced.
to rate and post a comment
Rick Latkovich
Re: Problem with AMX 4
June 20, 2008 03:50
The intermittent exposure problem can still be a contactor .Did you test contactors in diagnostic mode ? As far as your drive problem , you probabaly have aproblem with the tube either not sitting properly in cradle or the the sensor .If you have any problems with unit you can email your questions directly ,I will try to help out.
rlatkovich@alpha-imaging.com
to rate and post a comment
Rafael Cruz
Re: Problem with AMX 4
June 27, 2008 08:10
Stator speed.
to rate and post a comment
Rafael Cruz
Re: Problem with AMX 4
June 27, 2008 08:13
Sorry, forgot Question #2. Check if the bumper is free or the tube latch is ok.
to rate and post a comment