Jose Augusto
Problem with BENNETT HFQ 600 REF 380 66
September 16, 2009 09:48
I have a problem with a Bennett Hfq. By connecting the display shows REF 380 66 Can someone help me? jcarvalhojunior@globo.com
to rate and post a comment
JOSE ANTONIO SANCHEZ GALVAN
Re: Problem with BENNETT HFQ 600 REF 380 66
September 29, 2009 07:20
check the line voltage, perhaps there is a bad contact:.: or the circuit that mesures the input voltage is in bad condition .. jose a. sanchez g. josesangal@yahoo.com.mx
September 29, 2009 09:34
Bad contact in support of the fuse F6 ... Thank you for your attention. Jose
50+ Stretchers in stock for immediate rentals. Contact us for a today for a quote.
Jose Augusto
Problem with BENNETT HFQ 600 REF 380 66
September 16, 2009 09:48
I have a problem with a Bennett Hfq.
By connecting the display shows REF 380 66
Can someone help me?
jcarvalhojunior@globo.com
to rate and post a comment
JOSE ANTONIO SANCHEZ GALVAN
Re: Problem with BENNETT HFQ 600 REF 380 66
September 29, 2009 07:20
check the line voltage, perhaps there is a bad contact:.: or the circuit that mesures the input voltage is in bad condition ..
jose a. sanchez g.
josesangal@yahoo.com.mx
to rate and post a comment
Jose Augusto
Re: Problem with BENNETT HFQ 600 REF 380 66
September 29, 2009 09:34
Bad contact in support of the fuse F6 ...
Thank you for your attention.
Jose