Recently, I found an error "The x-ray controller is not generating A/C line interrupts. I have checked the emergency switch, x-ray enable key, incoming and scanner breaker side were okay. At last I checked the BOIS at the PC side and found the IRQ 10 was set to available. I change the IRQ 10 set to reserved and finally no error found & system operate normal.

But, after 2 to 3 days, I found the error was returned again and the IRQ 10 was set back to available.

I had checked the Motherboard battery is okay and I also try to change the PCI card. But after 2 week the error was came back again.

Anyone have the idea to solve the problems?
Or anything I need to check?

1 person liked this
Log inor Register

to rate and post a comment

(7)
(4)
(8)

Louie Arabo

re: Problem with Discovery-W

March 30, 2015 09:06

We can help you configure it. E-mail me at info@medpluses.com

Log inor Register

to rate and post a comment

(217)
(90)

MRAdmin

re: Problem with Discovery-W

April 04, 2015 10:13

it seems a problem with the motherboard, but if you did not change the battery you should do that even if it give 3 volt, it is best to replace it, and as a workaround for this error never disconnect the power to the PC.