Problem with Drager Babylog 8000

March 22, 2009 07:08

Hi, All friends
Babylog 8000 to the system was working well and two days ago is not working now .
When the system is working there is no pressure out from device .
Gives the errors on the screen :
( loss of stored data ) ( O2 pressure measurment inop ) .
I have tried a lot with him .
What is the solution .

Thanks,

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

Re: Problem with Drager Babylog 8000

March 23, 2009 10:09

If ypou have other equipment, try replacing pressure transducer for O2 or asociated components.
Discard ocludde filters
if you have not another equipment you need to identify the point where O2 pressure is taked and connect an DSO, then vary the O2 pressure (just connect and disconnect the hose from wall) if you can obtain a variation in the measure, the cpu board is the problem. if no read from DSO, then O2 pressure transducer may be the cause
Best Regards

Log inor Register

to rate and post a comment

(33)
(3)

Hassan Mohamed

Re: Problem with Drager Babylog 8000

March 24, 2009 02:10

Hi,
I change the oxygen sensitive and has not changed, nothing has been given the same system errors on the screen and does not leave the pressure of a final .

Thanks,

Log inor Register

to rate and post a comment

(31)
(2)

catur suharinto

Re: Problem with Drager Babylog 8000

March 31, 2009 06:31

may be the o2 pressure regulator is defective

Log inor Register

to rate and post a comment

(31)
(2)

catur suharinto

Re: Problem with Drager Babylog 8000

April 03, 2009 07:51

I have read the babylig service manual. the action for this corresponding error is Replace the O2 cell

Log inor Register

to rate and post a comment

(33)
(3)

Hassan Mohamed

Re: Problem with Drager Babylog 8000

April 04, 2009 02:50

Hi,
Thank you very much,
O2 have changed a new one has not changed anything.
The device works as well, even if O2 is either disabled or not there
I'm sorry if possible, send this service Manual to This email address is ( hmh2005_forever@hotmail.com ) .

Thanks,

Log inor Register

to rate and post a comment

(17)

jasmin dobariya

re: Problem with Drager Babylog 8000

April 06, 2015 08:01

hi...frnd
i havev also one babylog 8000 plus ventilator. and its given error 4.( and then given loss of stored data )
and not working good bcoz of this alarm .

is any solution for this Error??? give me some help for this error.

thanx in advance.. my frnd

Log inor Register

to rate and post a comment

(33)
(3)

Hassan Mohamed

re: re: Problem with Drager Babylog 8000

April 06, 2015 05:34

Hi, Jasmin
You need change the timekeeper .



Thanks,
Hassan

Log inor Register

to rate and post a comment

(17)

jasmin dobariya

re: re: re: Problem with Drager Babylog 8000

April 07, 2015 07:49

hello.
thanks for your reply.
but i have not any extra timekeeper. is this available in market??
do you have any timekeeper.????
give me your EMAIL ID so we can discuss more.
my Mail id is "jasdobariya@gmail.com"

thanks so much.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)