Problem with GE 600T
October 03, 2006 06:34
I am getting E130 and machine is not switching on.Which Pcb could be defective
to rate and post a comment
Ralph Childs
Re: Problem with GE 600T
October 03, 2006 09:53
Sunil,
Check the low voltage power supply, fuses and the incoming transformer. I had an exact problem like this about 6 months back and the transformer was sjorted.
Scooter
scanner1@eim.ae
to rate and post a comment
Manuel Casanova
Re: Problem with GE 600T
October 03, 2006 10:27
check The 400PL3 BOARD, and 400PL7 board, if you need more information contact me asap.