Problem with GE CT/e during scan
November 24, 2020 10:05
On GE CT/e scanner at the moment get scan (axial or helical) can not confirm AXIAEL SCAN ERROR Error 20-1001-26 Error detected by OGP DAS trigger of one rotation, between G-Pulse2 is under specifications on axial scan I already clean Slip rings and also exchange axial motor/encoder for another What can be wrong?
to rate and post a comment
KONSTANTINOS PATTICHIS
re: Problem with GE CT/e during scan
November 25, 2020 10:35
Measure the resistance between 6 test point on the board that touch the slipring and the back side of the sliprig . Must be about zero.
Matt Wojcik
November 30, 2020 10:35
Replace GPulse2 photo sensor
Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty! Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com
GEcase
Problem with GE CT/e during scan
November 24, 2020 10:05
On GE CT/e scanner
at the moment get scan (axial or helical)
can not confirm
AXIAEL SCAN ERROR
Error 20-1001-26 Error detected by OGP
DAS trigger of one rotation, between G-Pulse2 is under specifications on axial scan
I already clean Slip rings and also exchange axial motor/encoder for another
What can be wrong?
to rate and post a comment
KONSTANTINOS PATTICHIS
re: Problem with GE CT/e during scan
November 25, 2020 10:35
Measure the resistance between 6 test point on the board that touch the slipring and the back side of the sliprig . Must be about zero.
to rate and post a comment
Matt Wojcik
re: Problem with GE CT/e during scan
November 30, 2020 10:35
Replace GPulse2 photo sensor
to rate and post a comment