problem with ge voluson 730 pro
June 17, 2016 03:02
When I connect the probe in the screen don't show image.
to rate and post a comment
Pablo Delgado
re: problem with ge voluson 730 pro
June 17, 2016 06:06
Hi,
Did you try to test the probe in another system?
Did you get some kind of error?
The ultrasound system works with another probes?
to rate and post a comment
Lakshmi Narayana
re: problem with ge voluson 730 pro
June 19, 2016 02:59
It happens with all the probes? If so then your beamformer is faulty.
to rate and post a comment
EROTOKRITOS SKENTOS
re: problem with ge voluson 730 pro
June 19, 2016 02:59
If the system recognise the probe when you connect it but you do not have any image the first think you have to check is the power supply and then the BF board.
to rate and post a comment