Problem with image in advantx E RF system

Nacho hps

Problem with image in advantx E RF system

January 04, 2018 12:39

Dear I have a problem with the image in the table legacy when we make fluoroscopy the image go out like the image comes out distorted, like a sine wave in vertical and horizontal.
Well I check that the problem is coming from the II TH9447, I make the adjust of the tension with the program V004 image intensifier focus without result , also I change the configuration of the C007 I change the values of G1,and G2 and the problem persist in NORM and MAG1 and seems that in MAG 2 and MAG3 is much better and , but the problem in NORM and MAG1 continue is possible that have problems with configuration or maybe because was moved to another country initially it was working in some hospital in USA now is working in sud america Bolivia and the tension is 380 50 HZ.
Can you help me please

Log inor Register

to rate and post a comment

(78)
(7)

R GOPAL

re: Problem with image in advantx E RF system

January 05, 2018 09:25

I sincerely feel that the problem is not with the Image Intensifier. (The problem may be with the Camera ckt.)

First check the II output image viewing through the port. Confirm it is fine.

You should try and adjust manually G1, G2 etc to get the best image in norm and Mag. modes.

Now fix up the camera to the port and check.

R.Gopal