Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Problem with INTRA dental unit

Tags: Dental, PLANMECA

(2)

GARY MCROBERT

Problem with INTRA dental unit

June 23, 2020 10:29

I have a Planmeca INTRA dental unit with a technical problem I have the user, installation and parts manuals
none of these have information relevant to the problem
I am experiencing.

The unit appears to be functioning no error codes, control
panel reacts as expected except the unit fails to emit Xrays
when the expose button is pressed. The generator fails
to provide the high frequency high voltage drive to the
head. all fuses are intact the drive bridge has high voltage
green HV LED is illuminated, LED D9 illuminates during
expose button is activation.

I Suspect there is a communication protocol problem with
either the control panel, generator or the head. Each
component is of a different manufacture date by several
years perhaps there has been changes in the firmware.
can anyone help me solve this problem?

Thank you


Log inor Register

to rate and post a comment

(331)
(228)
(4)

Faustino Pereira

re: Problem with INTRA dental unit

June 24, 2020 10:47

You may have the demo program activated. It's just an idea, otherwise you would have error codes in the display

Log inor Register

to rate and post a comment

(2)

GARY MCROBERT

re: Problem with INTRA dental unit

June 25, 2020 10:32

Thank you for the reply. Can you instruct me, how to exit demo mode?

Log inor Register

to rate and post a comment