Milos Krstic

Problem with Philips SUPER 50 DO x-ray generator

April 14, 2009 08:53

During fluoroscopy there is not x-rays but also there aren't any errors. KV and mA on display are zero both in manual setting and automatic mode. I think that kontactors QK1 and QK2 in converter are not activated but what is the reason? I'm appreciated for any help.

Log inor Register

to rate and post a comment

(9)
(1)

Juan Martinez

Re: Problem with Philips SUPER 50 DO x-ray generator

April 18, 2009 09:28

Does the fluoro light on the desk comes on? Does it expose on small focal spot? Have you done Fluoro from the generator cabinet?

Log inor Register

to rate and post a comment

(47)
(4)

Milos Krstic

Re: Problem with Philips SUPER 50 DO x-ray generator

April 21, 2009 06:44

Super 50 DO is part of DORNIER HM3 lithotripter and it has two x-ray tubes and two TV chains. It's used only for fluoroscopy with both tubes. When I swiched on fluoro from remote hand switch got led and buzzer for x-rays,but really there is no x-rays. It's same for both tubes and for mannual and auto settings of kV and mA. I found that PRON signal is HIGH (should be low to activate conntactors QK1,2,3 for high voltage). I have only schematic diagrams without any explanation.
Also there is filament curent and it is changed if I mannualy change mA's.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

GE, Siemens, Philips, & Toshiba/Canon imaging parts, training & support!

DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)