Problem with Prestilix 1600S.

(25)
(5)

Roger Vanclooster

Problem with Prestilix 1600S.

September 19, 2011 01:44

I have the following problem with my Prestilix 1600S.
When I give the order for a movement of the selector to the right hand side, this movement starts with high speed and the table is cut off.
Movement to the left and the angulations are all normal.
Can anyone comment on this failure and give a hint to a solution ?
(The table has field modification instruction 11625A connected to print 1A38.)
R. Vanclooster.

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

Re: Problem with Prestilix 1600S.

September 19, 2011 04:58

Check the joystick ant the control desk, the reference pot is the same but the joystick could be defective. Have you try moving the table from the side of the table also?

Log inor Register

to rate and post a comment

(158)
(30)
(1)

tahar nouri

Re: Problem with Prestilix 1600S.

September 20, 2011 09:22

hi to all
i need service manual of MPG 50/60 xray generator

coupled to prestilix 1600

thankx in advance

Log inor Register

to rate and post a comment

(13)

cimon ching

Re: Problem with Prestilix 1600S.

September 19, 2011 09:48

your problem is quite common, you may check the control circuit with the relay control. I was away for this service long time ago, this was a know problem and a modification with it. check the motor relay control circuit with the transistor short circuited.
good luck for hunting.

cimon

Log inor Register

to rate and post a comment