Alexander Tonev

Problem with ProSpeed Qj

April 16, 2010 09:51

We have a Prospeed Qj. When we do warm up, there
is error ER-117 00000072 :Modification of Parameters is not allowed. I have no idea what that mean. Any solutions.
Thanks in advance.

Log inor Register

to rate and post a comment

(72)
(1)

Elliot Marcus

Re: Problem with ProSpeed Qj

April 18, 2010 07:36

My guess is that something in the generator is not being changed or being allowed to change. It could also be a problem with the tube or even system software.

Log inor Register

to rate and post a comment

(5)
(1)

zulquarnain shaikh

re: Re: Problem with ProSpeed Qj

November 27, 2013 04:42

I am also getting the same error msg with the same code. can you please tell me the tube of hispeed fxi is compatible with prospeed or not?

Log inor Register

to rate and post a comment

(73)
(8)

Kim Schofield

re: re: Re: Problem with ProSpeed Qj

November 28, 2013 08:16

The tubes are both MX-165 but not interchangeable.

Log inor Register

to rate and post a comment

(72)
(1)

Elliot Marcus

Re: Problem with ProSpeed Qj

April 18, 2010 07:37

If you can make exposures, varying the technique (kv and ma), without any errors and this only happens during a warm up, I would suspect that the problem may be with the system software. If you cannot, then your problem may be with the tube or generator.

Log inor Register

to rate and post a comment

Gilrubio

re: Problem with ProSpeed Qj

November 27, 2013 07:46

Try to reseat the boards in the electronic rack:
CPU, Kv Control, Etc.
Check the CTVRC for a blown IGBT

1 person liked this
Log inor Register

to rate and post a comment

(5)
(1)

zulquarnain shaikh

re: re: Problem with ProSpeed Qj

November 28, 2013 08:16

I have changes the whole erack and ctvrc too but the error remains the same? but one modification i have done is that i have placed tube of hispeed fxi in it, is that creat any problem?