Problem with Shimadzu EZy-Rad X ray
June 04, 2009 09:51
Hello
I'm using X ray EZy-Rad of Shimadzu
Some time machine occur error F31, If can pls send for me service manual
Pls advise for me this error
Thanks & best regards
to rate and post a comment
Swee Choon Neoh
Re: Problem with Shimadzu EZy-Rad X ray
November 04, 2009 09:53
The error sign showing that the current supply under voltage, it might fused one the fuses.
When you explore it with no image. I try to supply you operation manual
Dr. Swee Choon Neoh
to rate and post a comment
djamel amira
re: Re: Problem with Shimadzu EZy-Rad X ray
May 07, 2017 04:22
Hi;have you got service manual of this equipment ?if so please a copy ,i have f 22
to rate and post a comment
DMEDS
re: re: Re: Problem with Shimadzu EZy-Rad X ray
May 08, 2017 10:12
i have all service and schematic diagram for this system if you need it contact me,
magmesme@gmail.com
with best regards
to rate and post a comment
djamel amira
re: re: re: Re: Problem with Shimadzu EZy-Rad X ray
May 13, 2017 09:12
always nothing !
to rate and post a comment
DMEDS
re: re: Re: Problem with Shimadzu EZy-Rad X ray
May 08, 2017 10:13
error F22 mean
actually measured tube current is lower than set value by 50% or more.
regards
to rate and post a comment
DMEDS
re: re: Re: Problem with Shimadzu EZy-Rad X ray
May 08, 2017 10:13
F22 also mean,
failure of tube current control
Regards
to rate and post a comment
djamel amira
re: re: re: Re: Problem with Shimadzu EZy-Rad X ray
May 13, 2017 09:12
i need service manual how to make calibration of ma !
to rate and post a comment