We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.
rossana diaz
PROBLEM WITH TOSHIBA SSA-270A CODE ERROR
December 18, 2006 02:11
Im having this code error T/R ERROR TURN OFF This only happens when i connect the probes, i reseted the system and worked well, but as soon it gets hot , it turns out. Where can i start to llok the problem?
to rate and post a comment
Re: PROBLEM WITH TOSHIBA SSA-270A CODE ERROR
December 21, 2006 11:52
Does that happen with a specific probe ? check front end boards , probe connector , transmit power supply and the probe itself if it only happens with one probe.
to rate and post a comment
Dinh Hung
Re: PROBLEM WITH TOSHIBA SSA-270A CODE ERROR
December 22, 2006 12:23
Try to replace anorther probe, if the probe is OK, 99% the PULSER board has problem.
Dinh Hung
to rate and post a comment
Niranjan Kumar
Re: PROBLEM WITH TOSHIBA SSA-270A CODE ERROR
December 28, 2006 06:24
TR Error comes due to short of cables in the probe. Basically HV & Crystal wires. I think you have a problem with your probe.
to rate and post a comment