Problem withPolydoros 100 with cine attachment.

July 19, 2013 08:12

Dear All,

We have a Polydoros with cine attachment(Cath lab)
with PMT sensor.

Fluoroscopy works fine with good kV control according to the patient thickness proving that the PMT is funtional.

In cine mode, kV starts climbing and reaches to 125kV and remains there making the pictures very bright.

No error messages in the panel or console

Can anyone guide me to proceed further. I will be very much thankful