Prodigy Advance Error 22313: References values failure

Tags: Bone Densitometer, GE

(2)
(1)

Abel Salazar

Prodigy Advance Error 22313: References values failure

June 23, 2017 05:03

Hi, I get an equipment stopped working, it doesn't pass the daily QA. it does everything ok in the QA, but at the end displays failure. the error log shows: 22313 References values Failure.

I hope you can help me, thank you

1 person liked this
Log inor Register

to rate and post a comment

(12)
(1)

Mohammad Mahmoud

re: Prodigy Advance Error 22313: References values failure

June 25, 2017 05:19

Hi
Thee HV cables from both sides should seated good, also you can regrease using silicon oil .

Log inor Register

to rate and post a comment

(75)
(17)
(7)

Ben Williams

re: Prodigy Advance Error 22313: References values failure

June 27, 2017 10:30

Hi Abel,

Reference count failures are typically an indication of a failing x-ray tube. If you want to send us the complete error log I would be happy to take a look at it. It would be good to look and see if the tube is leaking any oil from the expansion bellows. That is fairly typical with this type of failure.

Ben
info@dexasolutions.com

Log inor Register

to rate and post a comment

This Comment has been deleted

June 27, 2017 10:31

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Don’t Compromise on Safety or Speed

From skilled GE depot repairs to patient-ready accessories, Southeastern Biomedical specializes in GE to keep your facility safe, efficient, and cost-effective. 888-310-7322.

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)