Hi, I get an equipment stopped working, it doesn't pass the daily QA. it does everything ok in the QA, but at the end displays failure. the error log shows: 22313 References values Failure.
Reference count failures are typically an indication of a failing x-ray tube. If you want to send us the complete error log I would be happy to take a look at it. It would be good to look and see if the tube is leaking any oil from the expansion bellows. That is fairly typical with this type of failure.
From skilled GE depot repairs to patient-ready accessories, Southeastern Biomedical specializes in GE to keep your facility safe, efficient, and cost-effective. 888-310-7322.
Abel Salazar
Prodigy Advance Error 22313: References values failure
June 23, 2017 05:03
Hi, I get an equipment stopped working, it doesn't pass the daily QA. it does everything ok in the QA, but at the end displays failure. the error log shows: 22313 References values Failure.
I hope you can help me, thank you
to rate and post a comment
Mohammad Mahmoud
re: Prodigy Advance Error 22313: References values failure
June 25, 2017 05:19
Hi
Thee HV cables from both sides should seated good, also you can regrease using silicon oil .
to rate and post a comment
Ben Williams
re: Prodigy Advance Error 22313: References values failure
June 27, 2017 10:30
Hi Abel,
Reference count failures are typically an indication of a failing x-ray tube. If you want to send us the complete error log I would be happy to take a look at it. It would be good to look and see if the tube is leaking any oil from the expansion bellows. That is fairly typical with this type of failure.
Ben