Johnny Barreto

Prospeed artifact issue

May 22, 2014 01:33

Dear Friends:

I am having a big headache troubleshooting this GE ProSpeed CT system, hilight detector. There are multiple rings or band, not sure how to explain, different shape of gray, and a spot in the middle of the image with a circle of another gray around it.
The CAM LINEARITY TEST, NO X-RAY, is all out of specification, but not only one or two pulses, neither a periodic pulses, but a curve with lot of pulses, some up of specs, some under the specs. Just replaced the X-ray tube, Bow tie filter and CIF board.

PLEASE HELP!!

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Prospeed artifact issue

May 22, 2014 06:50

was this happening before you replaced the tube?

Log inor Register

to rate and post a comment

(128)
(10)
(15)

Johnny Barreto

re: re: Prospeed artifact issue

May 23, 2014 10:04

yes, it was.

Log inor Register

to rate and post a comment

Mark Parris

re: Prospeed artifact issue

May 22, 2014 07:28

Look at the compensator inside the collimator. cracked

Log inor Register

to rate and post a comment

Mark Parris

re: Prospeed artifact issue

May 22, 2014 07:28

Any debris inside colinator?

Log inor Register

to rate and post a comment

(128)
(10)
(15)

Johnny Barreto

re: re: Prospeed artifact issue

May 23, 2014 10:04

I have cleaned collimator, tube, maask, new filter, check all alignment.

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: Prospeed artifact issue

May 23, 2014 07:29

Try reinstall a copy of cal tables, and check the power of DAS. Check matrix scan for stripes or lines witj no correction tables.

Log inor Register

to rate and post a comment

(128)
(10)
(15)

Johnny Barreto

re: re: Prospeed artifact issue

May 23, 2014 10:04

Dear Carlitos, please tell me where to find the procedures for checking matrix scan

Log inor Register

to rate and post a comment

(88)
(4)

Edurado Madeiros

re: Prospeed artifact issue

May 23, 2014 10:05

Let me know if the problem appears after changing tube , bow tie and Cif board? Or before that

Log inor Register

to rate and post a comment

(88)
(4)

Edurado Madeiros

re: Prospeed artifact issue

May 23, 2014 10:05

If You could, send me image to value. ermsismed@bol.com.br

Log inor Register

to rate and post a comment