ProSpeed ER-117 XG Error code 148
May 11, 2017 09:16
Hi guys, I'm having issues with this system, doesn't let me do the warm ups, I've checked all the interlocks and they are working just fine. ER-117 XG Error code 148 (Generator Safety Loop Open) Does any one knows where the Generator Safety Loop is located or what could be opening the loop?
to rate and post a comment
Lenny Place
re: ProSpeed ER-117 XG Error code 148
May 12, 2017 07:45
The generator safety loop if I remember comes in from the PDU to the strs assy. I would need to check the schematics which I have at home
Charbel Tadros
May 12, 2017 07:46
Thermal sensor of the tube . Check the wires.
John Giffen
May 13, 2017 09:09
Can you do an offline scan? Does the tube fire?
Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com
FranciscoG
ProSpeed ER-117 XG Error code 148
May 11, 2017 09:16
Hi guys, I'm having issues with this system, doesn't let me do the warm ups, I've checked all the interlocks and they are working just fine.
ER-117 XG Error code 148
(Generator Safety Loop Open)
Does any one knows where the Generator Safety Loop is located or what could be opening the loop?
to rate and post a comment
Lenny Place
re: ProSpeed ER-117 XG Error code 148
May 12, 2017 07:45
The generator safety loop if I remember comes in from the PDU to the strs assy.
I would need to check the schematics which I have at home
to rate and post a comment
Charbel Tadros
re: ProSpeed ER-117 XG Error code 148
May 12, 2017 07:46
Thermal sensor of the tube . Check the wires.
to rate and post a comment
John Giffen
re: ProSpeed ER-117 XG Error code 148
May 13, 2017 09:09
Can you do an offline scan? Does the tube fire?
to rate and post a comment