Hi guys, I'm having issues with this system, doesn't let me do the warm ups, I've checked all the interlocks and they are working just fine.

ER-117 XG Error code 148
(Generator Safety Loop Open)

Does any one knows where the Generator Safety Loop is located or what could be opening the loop?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: ProSpeed ER-117 XG Error code 148

May 12, 2017 07:45

The generator safety loop if I remember comes in from the PDU to the strs assy.
I would need to check the schematics which I have at home

Log inor Register

to rate and post a comment

(55)
(6)
(2)

Charbel Tadros

re: ProSpeed ER-117 XG Error code 148

May 12, 2017 07:46

Thermal sensor of the tube . Check the wires.

Log inor Register

to rate and post a comment

(2)

John Giffen

re: ProSpeed ER-117 XG Error code 148

May 13, 2017 09:09

Can you do an offline scan? Does the tube fire?