Over 500 Total Lots Up For Auction at One Location - NJ 06/15

ProSpeed II DBPCI#0 ERROR

Tags: CT, GE

Artemio Santacruz

ProSpeed II DBPCI#0 ERROR

January 26, 2014 08:00

Hello!!! every ones we are instaling a ge ct model ProSpeed II year 2008 when we turn on in power on test they show DBPCI #0 error
DBPCI #1 error
please help with your experience and kindnees
thanks and regard
Artemio Santacruz

Log inor Register

to rate and post a comment

(147)
(22)

Wang Zhixuan

re: ProSpeed II DBPCI#0 ERROR

January 27, 2014 08:43

The DBPCI is a board in a computer under OC,you should check the board. If you sure the board is defective, you
should change it. I can supply a good one to you.

Log inor Register

to rate and post a comment

Artemio Santacruz

re: re: ProSpeed II DBPCI#0 ERROR

February 04, 2014 08:03

thanks Wang fr answer me the problem was a acorrupted software and the solution a LFC
regards
Artemio

Log inor Register

to rate and post a comment

(67)
(7)

Galal

re: re: ProSpeed II DBPCI#0 ERROR

April 22, 2014 10:01

how we can know if it is defective ?

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS Mx-8000 IDT/Brilliance 16 Slice CT Scanner -----Worldwide leader!!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)