There are very few parts in the collimator that can be replaced. These were made by Siemens and adapted for GE mainly via the software on the chip set. I've never seen a service manual for the collimator itself, just calibration info. GE just swaps them out and Delta Medical an authorized dealer for Siemen's in our area does the same.
What is the problem you have? Maybe I can offer suggestions on what to do.
doug westlake
proteus collimator manual
August 10, 2012 01:07
does anyone have a manual for a proteus collimator
#alo1f??? this is a urgent request, please help.
to rate and post a comment
Pete Schliebner
re: proteus collimator manual
August 10, 2012 06:07
I have a parts manual but not a service manual. That would be handy.
to rate and post a comment
Staate Hayward
re: proteus collimator manual
August 13, 2012 10:54
There are very few parts in the collimator that can be replaced. These were made by Siemens and adapted for GE mainly via the software on the chip set. I've never seen a service manual for the collimator itself, just calibration info. GE just swaps them out and Delta Medical an authorized dealer for Siemen's in our area does the same.
What is the problem you have? Maybe I can offer suggestions on what to do.
to rate and post a comment
doug westlake
re: re: proteus collimator manual
August 13, 2012 11:51
cassette sizing is what is needed,i have done the fully open/closed cal but i believe there is more.
to rate and post a comment
Pete Schliebner
re: re: re: proteus collimator manual
August 13, 2012 03:03
My service manual for the Proteus system has a procedure in chapter 4. CAL 007 CALIBRATE CASSETTE SIZE DISPLAY 1 4-24.