Proview SNR dramatically low

October 06, 2010 09:39

Dear Friends.
i installed a USED proview MRI.
before dowing shim i tested the SNR of all the coils, they all was too low (about quarter of the norm for all)
does the bad shim may result in the a.m. trouble?
what could be the other posibilities?

regards

Log inor Register

to rate and post a comment

(28)

Veronica Alvarez

Re: Proview SNR dramatically low

November 03, 2010 11:03

Yes bad shim or leak in the rf room

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

Re: Proview SNR dramatically low

November 14, 2010 08:06

Well, A lot of possibilities.
1. Bad or no shimming after the installation.
2. Faulty shim (it can be tested using 2nd order shim calibration procedure) if calibration pass you can see if you get a maximum within the range (-1000 1000) for all the axis.
3. Bad RF calibration (after installation)
4. Incorrect phantom solution. (it is not a water)
5. Faulty Tx/Rx Pin driver
Please say more about your system installation process (shimming, calibration)