Abdelmuin Khatib

pulsera c-arm

February 11, 2016 04:25

Hello
I have c-arm model pulsera when I start it give me error #M 050 and M 400
I need help to solve problem
Regards

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: pulsera c-arm

February 11, 2016 05:54

you are getting the M050 because of the 400
M 400 IRIS X-ray disabled, Iris positioning error
Switch system off/on Call service if problem is persistent Iris positioning error on the collimator. X-ray is disabled. Check collimator functionality and cables. Perform Iris adjustment calibrations.

Log inor Register

to rate and post a comment

DMEDS

re: pulsera c-arm

February 12, 2016 09:01

050

System in
itialisation problem Switch system off/on. Call service if problem is persistent
A system startup problem occurred. Switch the system off and on. If the problem persists, check if the Mobile Viewing Station and the Stand are correctly connected. Check the Stand-Mobile Viewing Station cable.


400

X-ray disabled, iris positioning error. Call service
if problem is persistent

Iris positioning error on the collimator. X-ray is disabled. Check collimator functionality and cables. Perform Iris adjustment calibrations

dear sir if you need complete service manuals contact me,

magmesme@gmail.com

Reagrds

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)