Q4500 Error- Treadmill Power Failure

(4)

Scott Billingsley

Q4500 Error- Treadmill Power Failure

December 11, 2006 10:23

Q55 treadmill problem. Interface cable, control board, and speed sensor ruled out by sub. Swapped treadmills and problem follows treadmill. Cabled directly to control board with no solution. Driving me nuts!! Jeff sysmed@sysmed.org TIA

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

Re: Q4500 Error- Treadmill Power Failure

December 12, 2006 07:59

Jeff:
First, go away a couple of days. Let your mind take a breath...
Are you back?
Ok, Now change all the module, interface and spped sensor at the same time, probably you have a problem in both interface and module. Be sure the cables from speed sensor are ok, test it one by one and move it. Check mechanical obstruction of motor and poleis
I know the next look ridiculous but check the AC plug and ac cable, a loose cable can make this problem.

At last, a short circuit on motor can be the failure. You don't talk about to had tested it.

Try to be more specific about the fail: don't start, start and stop, noises?, etc.

If all fail, call Bewitched ;)

Good look

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Certified Pre-Owned Endoscopes, All Major Brands, Best Pricing & Warranty.

We are the best choice for all your Endoscopy needs. We have Certified Pre-Owned refurbished and used Endoscopes, and all other related components and accessories. We are committed to your absolute satisfaction. (513) 469-7267

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)