QA LUNAR DPXL Problem
May 08, 2007 10:36
After traslate a DPXL (uninstall C-Arm and then install) I stress steel cable for align xray tube with detector. All QA step pass except Spillover 9.75% (Before traslate Spillover 8.5%). I think that problem is connected with this align. Can somebody help me? What can I do?
Thanks
to rate and post a comment
ALEXANDER PILAFTSIS
Re: QA LUNAR DPXL Problem
May 09, 2007 12:44
When the spillover was 8.5%?
to rate and post a comment
Emerson Freire
Re: QA LUNAR DPXL Problem
May 09, 2007 01:05
Can you send me a copy o the QA printout?
But probably you may have a faulty detector...
emerson_freire@mandic.com.br
to rate and post a comment
Manuel Casanova
Re: QA LUNAR DPXL Problem
May 09, 2007 04:38
Dera:
Check the align betwen xray tube and detector, these is critical adjust.
If you need more options please contact m,e i live in Chile.
by
to rate and post a comment
Armando Peralta
Re: QA LUNAR DPXL Problem
May 09, 2007 08:27
Spillover was 8.5% before uninstall C Arm.
Thanks